Some measurements made to check how different modes affect the scope capture rate.

Oscilloscope basic specs
-3dB bandwidth: 1 GHz
Sample rate: 4 GSa/s (Ch1&Ch3 on); 2 GSa/s (all channels on)
Max Memory depth (real time mode): 2 Mpts (Ch1&Ch3 on); 1 Mpts (all channels on)
Min trigger re-arm time (segmented mode): 8 us
Claimed capture rate: 100 000 wfms/s

Note: Screen width is 10 div ==> "Max theoretical capture rate" = 1/(10x timebase)

Comments
Test 1: only Ch1
Test 2: Ch1 and Ch3 (same column as Test 1; no change in capture rate)
Test 3: Ch1 and Ch2
Test 4: All analog channels
Test 5: All analog + all digital channels
Test 6: All analog + all digital channels + decoding two 8bit buses
Test 7: All analog + all digital channels + decoding two 8bit buses, and zoom mode (zoom to 500 ps/div)
Test 8: All analog + all digital channels + decoding two 8bit buses, trigger on glitch > 45 ns
- 100k wfms/s was not acheived on the DUT in this setup; max measured was 94700 wfms/s
- Running two or four channels, adding 16 digital channels, decoding of two 8bit buses does not affect
the capture rate compared to single channel mode. On the contrary, enabling paired channels (Ch1&Ch2)
actually increases the capture rate! This is probably related to the lowered sampling rate.
- Trigging on "Glitch > 45 ns" does not affect the capture rate.
- Zoom mode decreases the capture rate at high timebase settings, by up to near 50%,  but
(surprisingly) increases the capture rate at low settings!
- Decoding of analog channels not tested.
- Display dot mode not tested

Used equipment
1) Function generator: Agilent 33522A
2) Timer/counter Stanford research systems SR620
Generator output terminated in 50 ohm.
Trig out terminated in 50 ohm.

Function generator setting:
10.000 MHz square wave, 50%, 1 Vpp
(10 MHz was chosen to be able to make the same test on scopes with higher capture rate.)

Counter connected to Trig out
Counter gate: 1s

Oscilloscope setting (unless otherwise noted)
ACQUSITION模式: normal
Horisontal mode: normal
X-position: mid
Trigger: edge trigger, rising edge, 500 mV, holdoff = min (60ns)
Display: vector mode


Capture rate [wfms/s] (rounded down to nearest 100 except for 10us/div --> )
Mode:     Ch1;      Ch1&Ch2   Ch1,Ch2,  +16 dig   + decode  enable    disable zoom,       Theoretical
          Ch1&Ch3             Ch3&Ch4             two 8bus  zoom      trig on             max capture
time-                                                                 glitch> 45 ns       rate [wfms/s]
base     
[ns/div)
0.5       93700     93600     93600     93500     93500     47000     93500               200 M

1.0       74100     93500     93500     93400     93400     47000     93400               100 M
2.0       74100     74000     74000     74000     74000     41500     74000                50 M
5.0       59900     59800     59800     59800     59800     36700     59800                20 M

10        59900     59800     59800     59700     59700     36600     59700                10 M
20        94700     59700     59700     59700     59700     36600     59700                 5 M
50        74200     73400     73400     73300     73300     47200     73200                 2 M

100       62400     53100     53100     53100     53100     42600     53100                 1 M
200       38700     33000     33000     33000     33000     32000     33000               500 k
500       16700     16100     16100     16100     16100     19400     16100               200 k

[us/div)
1         13300     11000     11000     1000      11000     14800     11000               100 k
2         10200     10600     10600     10600     10600     14700     10600                50 k
5          6300      9700      9700      6900      6900      7100     6900                 20 k

10                                                 3650                                    10 k
100                                                 780                                     1 k

[ms/div)
1                                                    60                                   100
10                                                    9                                    10
« 上次编辑:2013年1月13日,02:39:15 PM由Carlg »
 

在线的 EEVBLOG.

  • 行政人员
  • *****
  • 帖子:32206
  • 国家:  Au.
    • EEVBLOG.
RE:测试 - Agilent MSO7104B示波器bodog
« 回复#1开: 2013年1月09日,01:20:18 AM»
好工作,伟大的数据!

戴夫。
 

离线 卡尔格

  • 常规贡献者
  • *
  • 帖子:154
  • 国家:  SE.
RE:测试 - Agilent MSO7104B示波器bodog
« 回复#2开: 2013年1月13日,02:37:36 PM»
:phew: :phew: 它比我想象的客观测试更难。我从来没有学到的奇怪......

I 预期的 测试频率影响bodog。那'我选择了一个非常高的测试频率的原因,使其期间远低于范围's最小触摸/重新臂时间。我没有什么'期待是bodog会 减少 当我增加测试频率时 :o 它是一致的,在1MHz步骤中从1 MHz高达30 MHz。

然后我注意到即使改变测试信号的输出幅度也影响了bodog:较高幅度降低了捕获速率。接下来,我保持相同的幅度,但改变了范围垂直刻度,效果相同:从2 v / div到100 mv / div大大降低了bodog!是的,关闭所有通道给出最大捕获速率(其他设置不变)。

我假设只有捕获的数据量将产生效果(这是在这些安静的范围内自动),但显然's not the case. I'm不确定数据处理时间是否增加,或者它'S基本上更多的数据传输到图形处理器,降低了捕获速率。

结论是,当比较范围bodog时,似乎"fair"要使用自动设置,因为这么多参数出现。但是,自动设置结果可能不会(或者更常见)'t)是您可以选择典型测量的设置,这可能会提供"unfair"比较,就像在比较具有不同数量的水平分割的范围时。

此外,它可能值得注意的是 信号内容可能会改变捕获速率 在DSOS上,当高bodog很重要时,这可能很好。


在第一篇文章中进行一些添加(在所有设置上测试,因此可能存在异常):

ACQUSITION模式:
- 峰值检测没有改变bodog
- 平均模式和高分辨率模式降低了捕获速率

显示模式:
- 点显示模式没有改变捕获速率

水平模式:
- XY模式没有改变bodog
 